ESD safety in medical applications

Intona 7054-X(M), 7055-B, 7055-D
Post Reply
Daniel (Intona)
Posts: 104
Joined: May 6th, 2020, 8:29 am

ESD safety in medical applications

Post by Daniel (Intona) »

An always recurring question is how ESD safety can be achieved while maintaining isolation.

One thing is that it is quite impossible for a single device for both maintaining 4 or 5 kV isolation and having a path for ESD strikes through the isolated worlds. There are solutions for this, though. One solution is a combined EMI/ESD coupling capacitance that would be placed either at the isolation point or somewhere near to the human interface.

Another solution is risk management. You only need ESD protection at the particular points that can be touched by the user. And this is usually written in the product risk management policies, which are defined by you as manufacturer.

Here is a good article that helps to understand ESD in isolated environment:
[External Link Removed for Guests]
Daniel (Intona)

Daniel (Intona)
Posts: 104
Joined: May 6th, 2020, 8:29 am

Re: ESD safety in medical applications

Post by Daniel (Intona) »

And yes, we have an ESD gun in our lab and we love to help you to solve your issue.
Daniel (Intona)

Daniel (Intona)
Posts: 104
Joined: May 6th, 2020, 8:29 am

Re: ESD safety in medical applications

Post by Daniel (Intona) »

And don't worry, it's just ESD...
[External Link Removed for Guests]